Take the 2-minute tour ×
Biology Stack Exchange is a question and answer site for biology researchers, academics, and students. It's 100% free, no registration required.

I've been told that the maximum number of cycles in PCR is between 20 and 30.

Is this true, and what are the reasons for this limitation?

share|improve this question

2 Answers 2

up vote 8 down vote accepted

I would draw the line beyond 35, but thats a bit cosmetic. The reasons are manyfold:

  • due to the exponential fashion of the amplification (ideally) reagents are used up at some point
  • reagents degrade, this is especially true for the dNTPs
  • the activity of the enzyme, despite being heat-stable is declining over time
  • beyond 35 cycles the exponential curve is flattening out (reasons see above)
  • if you run the PCR for too long, you will get more and more side-products (mostly primer dimers, but mis-aligned primers can also make problems), this is more a problem for real-time PCR than for ones run on a gel

If you need a higher sensitivity with more cycles, you can use the technique called "nested PCR". There you do a first round with a primer pair specific for the region of interest and then do a second round with primers which are located slightly to the inside of the amplified DNA. This is done to avoid the amplification of unwanted contaminations. Since you do some 50-70 rounds of PCR amplification in total, this method is extremely sensitive (also to contaminations). See the image from the Wikipedia article for details:

enter image description here

share|improve this answer
1  
Thanks for the answer! Would it then be possible to terminate the PCR, isolate out all the DNA and then start a new batch in PCR with all the isolated DNA, as to renew the polymerase and have "new" primers? –  tschoppi Feb 14 '14 at 17:01
2  
Yes, thats possible and is done. But to avoid the amplification of a contamination, you use a set of primers which is located slightly to the inside of the product and which does not overlap with the primers of the first round. This technique is called nested PCR (the article also contains an image of the technique. –  Chris Feb 14 '14 at 17:14
    
Awesome, thanks for the info! –  tschoppi Feb 14 '14 at 17:19
    
I have added it to the answer. –  Chris Feb 14 '14 at 17:23

How many cycles of PCR before dNTPs run out?

Assume a 25 μl reaction.

Assume 200 μM dNTPs.

200 μM dNTPs = 200 pmol μl -1

so in 25 μl reaction, there are 5000 pmol of dNTPs

= 5000 x 10-12 x 6 x 1023 molecules

= 3 x 1015 molecules dNTP

Assume that we start with 1 molecule of a 1000 bp template, 50% GC

1 kb = 2000 nucleotides

So , how many of these molecules can we construct using 3 x 1015 molecules dNTP?

= 3 x 1015/2000

= 1.5 x 1012 DNA molecules

How many cycles of PCR to produce this many from a single template molecule?

2n = 1.5x 1012

nlog2 = log(1.5) + log(1012)

nlog2 = 12.18

n = 12.18/ log2 = 41 cycles

Of course this in an absolute upper limit. The estimate assumes that you start with one template molecule of 1 kb, and that dNTPs aren't being hydrolysed, or otherwise degraded.

share|improve this answer
    
You have to take into account, that we use four different dNTPs for PCR, so this calculation would only be valid if the bases are equally distributed. –  Chris Feb 14 '14 at 18:44
1  
Which is why I stated the assumption of 50% GC (and so 50% AT). –  Alan Boyd Feb 14 '14 at 21:27
    
You are right, I missed that point. Sorry for that. –  Chris Feb 14 '14 at 21:37

Your Answer

 
discard

By posting your answer, you agree to the privacy policy and terms of service.

Not the answer you're looking for? Browse other questions tagged or ask your own question.