$\begingroup$

I ran a PCR product of ~300 bp on a 2% TAE-agarose gel for 30 minutes. I used Sybr-safe as a DNA stain. Voltage was 80V.

When I imaged the gel, the DNA on the bottom half of the gel, including the ladder had disappeared (it showed no bands). The DNA in the top half looked-clear and well-separated, but the bands on the bottom were somehow missing.

What could be the reason? I've never encountered this before.

I know it's not due to gel overheating - I ran the gel in the cold room and when it was done it was cold to the touch.

$\endgroup$
1
  • $\begingroup$ Just stain your gel again for 10 min protected from light with slight agiation. DNA bands will appear again. Try! $\endgroup$
    – user4325
    Aug 26, 2013 at 9:06

1 Answer 1

3
$\begingroup$

If you use pre-staining, the stain migrates as well in the electric field and this can lead to some parts of the gel being unstained. This would look pretty much like you describe, the ladder being invisible as well points strongly to that part of the gel just being no stained properly.

It sounds a bit unlikely with your running parameters, I've only seen that with a very long running gel, but I also haven't used pre-staining for quite a while now. Ethidium bromide is positively charged and migrates in the opposite direction as the DNA, I have no experience with Sybr-Safe though and don't know if it behaves the same as EtBr in this regard.

$\endgroup$
4
  • $\begingroup$ I looked it up and the website does say sybr safe runs away from migrating DNA. I might have been off with the timing and this fits, thanks. $\endgroup$
    – Drosophila
    Apr 4, 2013 at 12:47
  • $\begingroup$ @Drosophila I'd just try to stain the gel again, if this was the reason the DNA should be visible then. $\endgroup$ Apr 4, 2013 at 12:49
  • 1
    $\begingroup$ Yeah, I threw out the old gel but I ran a new one at a higher voltage and shorter time. It worked. I'll check your theory when I get some free time $\endgroup$
    – Drosophila
    Apr 4, 2013 at 16:05
  • $\begingroup$ Jut for reference, with 20min@100V running time I usually get the bottom ~1.5cm of the gel Sybr-Safe free. $\endgroup$
    – Armatus
    Feb 9, 2018 at 11:20